[Date Prev][Date Next][Thread Prev][Thread Next][Date Index][Thread Index]

Re: A proposal for setting a BCAL threshold




Elke,

Right -- but it looks like the rate for double PE equivalent hits is  
down by an order of magnitude.  So it should be about a 10%  
correction.  I think at the level we are trying to estimate this I'm  
happy to operate on the assumption that the single PE rate completely  
dominates the dark rate.

Cheers,

-Matt

On Jun 8, 2007, at 4:00 PM, Elke-Caroline Aschenauer wrote:

> On Fri, 8 Jun 2007, Matthew Shepherd wrote:
>
> Matt,
>
> just one remark the assumption that all of the dark rate will be one
> photon electron is not really true for SiPMTs, please look to
> /home/elke/public/070419_jlabs_presentation.pdf page 29 and 30
> as a pixel can only be on or off, contrary to other devices.
>
> bye elke
>
>
>> Date: Fri, 8 Jun 2007 15:38:09 -0400
>> From: Matthew Shepherd <mashephe@indiana.edu>
>> To: George Lolos <George.Lolos@uregina.ca>
>> Cc: Richard Jones <richard.t.jones@uconn.edu>, halld-cal@jlab.org
>> Subject: Re: A proposal for setting a BCAL threshold
>>
>>
>> Richard and George,
>>
>> Thanks for the info.  So 32 MHz was maybe a bit optimistic -- let's
>> take then 60 MHz.  At 60 MHz, the probability to have 11 or more dark
>> pulses in a 100 ns window is 0.04 (less than 5%).  So I will bump the
>> threshold up to 11 photoelectrons or about 2.4 MeV.
>>
>> I propose a 1 MeV threshold coming out of HDGeant.  Then in the BCAL
>> response code in DANA I'll introduce some sampling fluctuations and
>> make a second cut at 2.4 MeV.
>>
>> -Matt
>>
>>
>> On Jun 8, 2007, at 1:12 PM, George Lolos wrote:
>>
>>> Hi Richard and Matt:
>>>
>>> We had a phone conference yesterday with SensL and they have just
>>> tested the wafers for the arrays under Phase 1.
>>> The 820L pixel type has demonstrated the following performance at
>>> room temperature:
>>>
>>> Voltage above breakdown                         Pure PDE (no cross
>>> talk or after pulsing)                DR/pixel                  DR/
>>> Array
>>>
>>>           +
>>> 1V
>>>     7%                                                   800
>>> Hz                  65 MHz
>>>           +
>>> 2V
>>>   11%                                                 1700
>>> Hz                137 MHz
>>>           +
>>> 3V
>>>   13%                                                 3200 Hz
>>>           +
>>> 4V
>>>   17%
>>> At a temperature of +5 C  the dark rates are cut by a factor of
>>> 2.6, so without an elaborate or expensive cooling we can bring this
>>> down to around 53 MHz at 11% PDE.   These then will be for Phase
>>> 1.  For Phase 2, the new Si process they have just implemented has
>>> reduced the DR by a factor of just over 2 and they expect to
>>> further improve and come down to that of the original Si treatment
>>> that does not give good reproducibility of breakdown voltage.
>>> Trenching will also reduce cross talk and keep the DR at the 1 P.E.
>>> level.
>>>
>>> The bottom line is that Richard is correct that at the present time
>>> 32 MHz at PDE values larger than have not been achieved.   Are such
>>> rates achievable for production?  I believe they are but aiming at
>>> a more realistic 50-60 MHz at PDE ~ 10-15% is the wiser route to go.
>>>
>>> I hope this helps,
>>>
>>> George
>>>
>>>
>>> Richard Jones wrote:
>>>
>>>> Matt,
>>>>
>>>> The design goal of 32MHz may eventually be achieved, but this is
>>>> not demonstrated.  At the last meeting George agreed that 100MHz
>>>> is achievable with what he has seen.  I would believe something
>>>> more like 150MHz @ 22C from what I have actually seen, but I am
>>>> not working with Sensl modules.  This depends a lot on
>>>> temperature, but we agreed that we do not want to have to
>>>> refrigerate the BCal readout very much.
>>>>
>>>> Richard J.
>>>>
>>>> Matthew Shepherd wrote:
>>>>
>>>>>
>>>>> Hi All,
>>>>>
>>>>> Here's a proposal for setting a BCAL threshold so we can start to
>>>>> refine the reconstruction a little bit.
>>>>>
>>>>> - take dark rate at 32 MHz (design goal from GlueX-doc-795) and
>>>>> assume this is only single PE rate
>>>>> - for a 100 ns window this means an average of 3.2 pulses per  
>>>>> window
>>>>> - assume the fADC processing just generates a pedestal subtracted
>>>>> mean and that dark rate (not electronics noise) dominates the
>>>>> pedestal
>>>>> - let's assume the DAQ can handle 5% occupancy in the BCAL
>>>>> - if average is 3.2 dark pulses, the probability of having 7 or
>>>>> more pulses in a window is 0.04
>>>>>
>>>>> --->> set threshold at 7 photoelectrons
>>>>>
>>>>> 7 photoelectrons * ( 26 keV_fiber / pe ) / 12% = 1.5 MeV energy
>>>>> deposited in cell
>>>>>
>>>>> I propose we adjust the threshold to 1.5 MeV (down from 10 MeV)
>>>>> and work from there.  Of course this needs further study, and
>>>>> validation through whatever bench studies, beam test, etc. etc..
>>>>> My main goal is to get around the right order of magnitude so we
>>>>> can make another pass at the reconstruction algorithm which will
>>>>> behave very differently with this much lower threshold.  Does
>>>>> anyone see a serious flaw with this?
>>>>>
>>>>> -Matt
>>>>>
>>>>
>>>
>>>
>>
>>
>
>  ( `,_' )+=-+=-+=-+=-+=-+=-+=-+=-+=-+=-+=-+=-+=-+=-+=-+=-+=
>   )    `\                                                  -
>  /    '. |                                                  +
>  |       `,              Elke-Caroline Aschenauer            =
>   \,_  `-/                                                    -
>   ,&&&&&V         Jefferson Lab                                +
>  ,&&&&&&&&:       HALL-D 12C / F381       121-A Atlantic Avenue =
> ,&&&&&&&&&&;      Mailstop: 12H5          Hampton, VA 23664      -
> |  |&&&&&&&;\     12000 Jefferson Ave                             +
> |  |       :_) _  Newport News, VA 23606  Tel.:  001-757-224-1216  =
> |  |       ;--' | Mail:  elke@jlab.org    Mobil: 001-757-256-5224   -
> '--'   `-.--.   |                                                    +
>    \_    |  |---' Tel.:   
> 001-757-269-5352                             =
>      `-._\__/     Fax.:   
> 001-757-269-6248                              -
>             +=-+=-+=-+=-+=-+=-+=-+=-+=-+=-+=-+=-+=-+=-+=-+=-+=-+=- 
> +=-+=-+
>