[Date Prev][Date Next][Thread Prev][Thread Next][Date Index][Thread Index]

Re: CDC upstream end config



Hi Curtis,

I believe I didn't fully address the first part of your question on your 
earlier email. This has to do with the use of the shielded cable 
connecting the straws to the connectors.

On option #1, a voltage trip  turns off the HV supply for a HVB card (24 
straws). In this case, the cable does not need to have a spec meeting 
the operating voltage of the straws because the potential difference 
between the inner conductor and the shield (connected to the same HV 
supply channel) is always about zero. The cable shields need to be 
insulated to prevent contact with adjacent groups of 24 straws. The 
cable can be very thin.

On option #2, there is a requirement to disable individual straws within 
the same group of 24 straws. In this case, the shielded cable needs to 
meet the operating HV rating at all times. The shield also needs to be 
connected to a fixed, low impedance reference, either HV or ground 
(attach to the connector accordingly). Connecting to ground has the 
advantage that the shield does not require an insulation jacket. The 
cable cannot be as thin as the one in option #1.

Please refer to my earlier email regarding the cable specs.

Regards,
Fernando



Fernando J. Barbosa wrote:
> Hi Curtis,
>
> Coax cable manufacturers specify the operating voltage between the  
> inner conductor and the shield and where the shield is normally at 
> ground - most applications. This spec refers to the dielectric between 
> the inner conductor and the shield, among other variables. The 
> manufacturers also specify the type and dimensions of the jacket 
> insulating material which can then be used to determine their usage or 
> application.
>
> Regards,
> Fernando
>
>
>
> Curtis A. Meyer wrote:
>> Hi Fernando -
>>
>>    I know that you have stated that the shield needs to be at HV, but 
>> I originally thought this was because the cable you had was not 
>> spec'd for high voltage. Are we sure that any cable is really spec'd 
>> to have high voltage on the shield? Most manufacturers spec the inner 
>> wire relative the the shield and the outside world, but I don't 
>> recall seeing
>> a spec from the shield to the outside world?
>>
>>     Curtis
>>
>> On Mon December 8 2008, Tim Whitlatch wrote:
>>  
>>> Thank you Fernando for your reply. This would require someone to 
>>> install the FEP jacket on all the wire? Is this shrink wrap?
>>>
>>> Thanks,
>>>
>>> Tim
>>>
>>> Fernando J. Barbosa wrote:
>>>    
>>>> Hi Tim,
>>>>
>>>> My comments follow:
>>>>
>>>> Tim Whitlatch wrote:
>>>>      
>>>>> Hello All,
>>>>>
>>>>> Per our video conference this morning, the following was discussed;
>>>>>
>>>>> Gary will continue to manufacture a 24 place upstream plate with 
>>>>> proper geometries from "geometry C". this will give room for 
>>>>> simulating worst case spacial considerations for assy. He will 
>>>>> pursue the original type setup with 5.5 inch long cables max.
>>>>>
>>>>> As a backup, Gary will manufacture an offset plate with the 24 
>>>>> holes to take the stress off the conductive rubber sleeve (per SLAVA)
>>>>>
>>>>> Fernando, Gary followed up on the coax cables you suggested. They 
>>>>> all seem to come without an outside insulating jacket. Is it 
>>>>> possible to use these cables by keeping the braided shield at 
>>>>> ground? Also, is it possible to use cable that is not rated for 
>>>>> our high voltage by doing some tests?
>>>>>         
>>>> Of the two suggestions to check the cable routing, one of the 
>>>> coaxes should have a jacket but the semi-rigid cable should have an 
>>>> aluminized jacket. In this case, an insulating sleeve/jacket needs 
>>>> to be installed and there are various types offered by multiple 
>>>> companies. I suggest FEP, if available in the required diameter. 
>>>> The shield needs to be at HV and in normal operation the potential 
>>>> difference between the shield and the inner conductor is zero. 
>>>> However, if a straw wire is disabled, the full HV potential will be 
>>>> present and the cable needs to conform to these requirements.
>>>>
>>>> Cables can certainly be tested at higher voltages than the rated 
>>>> operating voltage specified by the manufacturer but that's a safety 
>>>> issue. I strongly recommend against overlooking the manufacturer's 
>>>> safety specifications.
>>>>
>>>> Regards,
>>>> Fernando
>>>>
>>>>
>>>>      
>>>>> I will pursue with Tom Carstens from Hall B about connector leaks 
>>>>> on their drift chambers.
>>>>>
>>>>> Yves will look at getting some extension pcbs made for testing 
>>>>> Slava's ides per layout on 9-29-08. Can we get a connector made 
>>>>> directly onto the HV board that would plug into this?
>>>>>
>>>>> Thanks for your time,
>>>>>
>>>>> Tim
>>>>>
>>>>>
>>>>>
>>>>>
>>>>>         
>>
>>
>>
>>   