Hi Elton, I hope this answers everyone's questions about the time resolution and any factors of two. I have told you the formula I have been using incorrectly. The short note I've attached should clear this up. -Blake Elton Smith wrote: Pine.LNX.4.58.0802130755590.5640@elton01.jlab.org">Hi Blake, Thanks for looking into this. But I think the equation for z should not have the 1/2 for this new definition because of the energy terms: DeltaT ={t_north*E_north - t_south*E_south}/Sum{E_north+E_south} At least for the case where E_north - E_south, then DeltaT = 1/2 (t_north - t_south) z = veff * (Tleft - Tright)/2 = veff * DeltaT Or am I still missing a factor? Thanks, Elton. Elton Smith Jefferson Lab MS 12H5 12000 Jefferson Ave Suite # 16 Newport News, VA 23606 elton@jlab.org (757) 269-7625 (757) 269-6331 fax On Tue, 12 Feb 2008, Blake Leverington wrote:Hi Elton, I started rerunning the timing resolution code this morning so hopefully it should be done tonight. I removed the 1/2 from the previous DeltaT so that DeltaT now is DeltaT ={t_north*E_north - t_south*E_south}/Sum{E_north+E_south}, without the factor of 1/2. I'll see how this affects the resolution but I suspect it will double it as z was previously assumed to be z = 0.5 *veff *DeltaT so that the z resolution should be double, from 0.7cm to 1.4cm at 1 GeV. I'll even redo the mean time resolution tomorrow as I've thought of new things to do. -Blake Elton Smith wrote:HI Blake, There is still another factor of two. If you simply remove the 1/2 in the note for E_N=E_S, then DeltaT = 1/2 delta_T and one also needs to remove the 1/2 in the equation for z. Now, removing the 1/2 in the note is fine as long as that is what you actually did to obtain the sigma. And this should be checked. Otherwise, the measured width is either higher or lower relative the actual value of delta_T which is what we believe we are measuring. So can you check what was actually calculated to obtain the width? Thanks, Elton. Elton Smith Jefferson Lab MS 12H5 12000 Jefferson Ave Suite # 16 Newport News, VA 23606 elton@jlab.org (757) 269-7625 (757) 269-6331 fax On Tue, 12 Feb 2008, Blake Leverington wrote:Hi Elton, I was just reviewing my notes, and it seems that if you assume that z = veff * delta_T / 2 then there should NOT be the factor of 0.5 in DeltaT. DeltaT should equal delta_T for E_N = E_S. I should remove that factor of 1/2 in the note. -Blake Elton Smith wrote:Hi Blake, Yesterday we were going over the correspondence between the expected z position resolution and the time resolution. We have the following relation: z = veff * (Tleft - Tright)/2 delta_z = veff * delta_T/2 for delta_T = Tleft - Tright We have been assuming that the quoted delta_T (75 ps/sqrt(E) + 30) is given by the definition above. But the reference for the resolution is GlueX-doc-804 which you wrote. Refering to eqs 14, you use the following expression: DeltaT = 0.5*Sum{t_north*E_north - t_south*E_south}/Sum{E_north+E_south} At first, it looks like the definition for DeltaT = 0.5*delta_T, removing the 1/2 from the equation for z. However, on closer inspection, there are more questions. Taking the simple case where E_north=E_south (center of calorimeter), then DeltaT (E_north=E_south) = 0.25 * (t_north - t_south), so it looks like there is even an additional factor of 2. Can you help me sort this out? Thanks, Elton. Elton Smith Jefferson Lab MS 12H5 12000 Jefferson Ave Suite # 16 Newport News, VA 23606 elton@jlab.org (757) 269-7625 (757) 269-6331 fax |